We evaluated one of these recently and found terrible correlations between
the Chiron 855 and the IRMA for Na, Ca+, and especially pO2 - apparently a
problem with sample application not pushing the calibration gel completely
off the electrodes. The precision for the pCO2 was also poor from aqueous
QC. I'm not convinced that a limited number of users would get enough
practice with the injection technique to get reliable results especially for
the pO2.
There are 1 or 2 abstracts/papers confirming this. The large wastage of
cartridges repeating the analyses will also make this analyser very
expensive to run.
Dr Fiona Ivison
Dept. of Clinical Chemistry
Royal Liverpool University Hospital
4th Floor Duncan Building
Liverpool
L7 8XP
0151 706 4287