This weeks Lancet article on Troponin (Vol 354; 1757-62) showed that the 
AxSYM troponin I assay appeared to have a useful clinical utility in the 
PRISM study.
The AxSYM cut-off of 1ug/L appeared to distinguish the patients with acute 
coronary syndromes who were likely to benefit from 
glycoprotein-IIb/IIIa-receptor antagonists from those who were not.
Perhaps this is the best evidence-based level to use.

Dr. Eric S. Kilpatrick
Consultant in Chemical Pathology
Department of Clinical Biochemistry
Hull Royal Infirmary
Anlaby Road
Hull HU3 2JZ

Tel 01482-674312
Fax 01482-674310


>Regarding cuttoffs, we are guided very heavily by the literature, and there
>has been remarkably little peer-reviewed data using the AxSYM so far.
>Abbott are getting there and the best source of information is the package
>insert.
>So: we use their disease-free reference interval of </= 0.5 ug/L (most
>walking-well are well below this) and a diagnostic cuttoff for AMI of 2.0
>ug/L. The unanswered question to me at present is what level is significant
>with regard to Minor Mycardial Damage (MMD) with associated poor prognosis
>in the UAP patient group? None of the Abbott data addresses this problem.
>Causes of elevated troponin (>0.5 ug/L) on the AxSYM then can be due to:
>AMI (usually above 2.0 ug/L, can reach several hundred ug/L); MMD -
>indicating the need for inpatient management (obviously moving the cuttoff
>up from 0.5 ug/L will improve specificity for coronary disease / cardiac
>event prediction but at a loss of sensitivity - I have seen little data eg
>ROC curve on the Abbott assay for this - Hans, please publish soon!); renal
>failure (5% above 1.2 ug/L - Abbott); Skeletal muscle damage (5% above 1.0
>ug/L - Abbott); cardiac bypass surgery (gross elevations peak 3-6 hours
>post-op - in-house data); peri/myocarditis; cardiac trauma; other cardiac
>procedures (eg stenting).
>
>Experience at at least one Hospital in Sydney was to discontinue the use of
>the assay due to too many "false-positives" leading to overcrowding of
>coronary care beds. Abbott have raised the cuttoff from 0.4 to 0.5 since
>then and the hospitals troubles were probably exacerbated by the well-known
>episode of quality problems with the assay. I think (unsubstantiated
>opinion) that a cuttoff of about 0.6-0.8 ug/L would give a suitable 
>balance.
>
>As an aside I hope to see vast improvements in TnI assays in the near
>future. The Dade Stratus method shows an order of magnitude improvement in
>precision and this looks like it will allow much better patient
>stratification in the troublesome range (Clin Chem 1999;10:1789-96).
>
>These are my opinions and should be treated as such. I would be interested
>in any other thoughts on cuttoffs for different troponin assays.
>
>Best wishes,
>
>
>
>
>Graham Jones
>
>Staff Specialist in Chemical Pathology
>St Vincent's Hospital, Sydney