not being able to break into the machine as such, i could only read what 
is in the DICOM header and therefore could be changed at the console. it 
says that the 'flip angle' is 180 degrees.
s



Peter Kochunov wrote:
> No, I don't think so. That would have been true under "classic CPMG" 
> conditions, however, it won't work for siemens MSE sequence. Could you 
> check on the angle value of the refocusing pulse? If you haven't 
> changed it, it is probably still 120 degrees e.g. the value that is 
> the siemens template.
> pk
>
>
> ----- Original Message ----- From: "Scott Kolbe" 
> <[log in to unmask]>
> To: <[log in to unmask]>
> Sent: Monday, May 05, 2008 8:45 PM
> Subject: Re: [FSL] multiecho T2 image acquisition question
>
>
>> Peter,
>> Do you think it is valid to use the CSF decay in the even echos to 
>> correct the odd echos? would the odd echo images be affected 
>> linearly? i am really hoping to be able to use the first echo to try 
>> and calculate myelin water fraction.
>> s
>>
>> Peter Kochunov wrote:
>>> Depends on the sequence some deal with that better than others. The 
>>> stimulated echo  contributes to the second echo and then it is being 
>>> refocused with the main echo, the contributions from the stimulated 
>>> echos from the third, and later pulses is going to be smaller since 
>>> magnetization is going to decay, but it can produce an odd-even echo 
>>> differences you observe. And yes, siemens MSE sequence isn't 
>>> particularly good as far as dealing with stimulated echo. For this 
>>> sort of sequence, I think the most robust processing will be to 
>>> discard the first echo and just fit the rest of the data.
>>> pk
>>>
>>> ----- Original Message ----- From: "Scott Kolbe" 
>>> <[log in to unmask]>
>>> To: <[log in to unmask]>
>>> Sent: Monday, May 05, 2008 8:22 PM
>>> Subject: Re: [FSL] multiecho T2 image acquisition question
>>>
>>>
>>>> Thanks Peter, Steve and David for the insights.
>>>> One follow-up question, would the stimulated echo effect be 
>>>> expected to dimish in later odd echos? Here is are the intensity 
>>>> values for a voxel in ventriclar CSF which one wouldn't expect to 
>>>> vary much in 132ms:
>>>> 1333
>>>> 1567
>>>> 1487
>>>> 1509
>>>> 1490
>>>> 1479
>>>> 1473
>>>> 1453
>>>>
>>>> As you can see, the odd/even difference is apparent but it is small 
>>>> compared to the difference between the first and second echos. Is 
>>>> it possible that the first refocussing pulse could vary in accuracy 
>>>> to later ones or that the spoiler gradients could vary in 
>>>> magnitude? This sequence was acquired on a 3T Trio TIM if anyone 
>>>> has experience with this kind of imaging on this particular machine.
>>>>
>>>> Thanks again for your time,
>>>> Scott
>>>>
>>>>
>>>> Peter Kochunov wrote:
>>>>> Scott,
>>>>> To be very technical, it is the stimulated echo produced by your 
>>>>> second refocusing pulse, due to less then perfect refocusing, that 
>>>>> adds the extra signal to the even echoes. So, one way to deal with 
>>>>> this to fit your data based on even and then odd echoes and 
>>>>> average the results. There are ways of reducing your stimulated 
>>>>> echo, first make sure that you do use 180 degree pulses for 
>>>>> refocusing. You can try changing the refocusing pulse for an 
>>>>> adiabatic. And if you have access to the pulse sequence you can up 
>>>>> the magnitude of the crusher gradients, or even make them asymmetric.
>>>>> pk
>>>>>
>>>>> ----- Original Message ----- From: "David Lythgoe" 
>>>>> <[log in to unmask]>
>>>>> To: <[log in to unmask]>
>>>>> Sent: Monday, May 05, 2008 8:29 AM
>>>>> Subject: Re: [FSL] multiecho T2 image acquisition question
>>>>>
>>>>>
>>>>>> Hi Scott,
>>>>>>
>>>>>> In multiple spin echoes, errors in the flip angle for the 
>>>>>> refocusing pulses
>>>>>> can lead to reduced signal on odd echoes. These are compensated 
>>>>>> for to some
>>>>>> extent on even echoes.
>>>>>> Even if you get perfect 180 pulses at the centre of your image 
>>>>>> slices, the
>>>>>> slice profile won't be perfectly square, and you will have 
>>>>>> reduced flip
>>>>>> angle further from the centre of the slice.
>>>>>> Do all your even echoes have higher signal than expected?
>>>>>> Also, remember the T2 for CSF is quite long.
>>>>>>
>>>>>>
>>>>>> Dave
>>>>>>
>>>>>>
>>>>>>
>>>>>> On 5/5/08 10:05, "Scott Kolbe" <[log in to unmask]> wrote:
>>>>>>
>>>>>>> Dear FSL list
>>>>>>> this is a question not directly related to FSL but I am a bit 
>>>>>>> lost for
>>>>>>> an alternative place to ask for advice. I have acquired 
>>>>>>> multiecho T2
>>>>>>> images to study relaxation times in a patient group. i found 
>>>>>>> that for
>>>>>>> the first echo image (13ms) the CSF signal is actually darker 
>>>>>>> than in
>>>>>>> the second echo image (26ms) by about 10%. i have acquired 8 
>>>>>>> echos and
>>>>>>> for the final echo image (132ms), the CSF signal is only 3% 
>>>>>>> darker than
>>>>>>> for the second echo.
>>>>>>> My question is, does anyone have any idea why the first echo 
>>>>>>> would be
>>>>>>> darker than the second. i had presumed that the earlier the echo is
>>>>>>> collected, the more signal there is. is that true in practice?
>>>>>>>
>>>>>>> thanks in advance
>>>>>>> Scott
>>>>>>>
>>>>>>> ========================
>>>>>>> Scott Kolbe
>>>>>>> Postgraduate Student
>>>>>>> Neuroimaging Group
>>>>>>> Howard Florey Institute &
>>>>>>> Centre for Neuroscience
>>>>>>> University of Melbourne
>>>>>>> VIC, Australia, 3010.
>>>>>>>
>>>>>>> ph:       +61 3 8344 1887
>>>>>>> email:    [log in to unmask]
>>>>>>> website:  http://www.neuroimaging.org.au/index.php?id=383
>>>>>>
>>>>
>>>>
>>>> -- 
>>>> ========================
>>>> Scott Kolbe
>>>> Postgraduate Student
>>>> Neuroimaging Group
>>>> Howard Florey Institute &
>>>> Centre for Neuroscience
>>>> University of Melbourne
>>>> VIC, Australia, 3010.
>>>>
>>>> ph:       +61 3 8344 1887
>>>> email:    [log in to unmask]
>>>> website:  http://www.neuroimaging.org.au/index.php?id=383
>>>>
>>
>>
>> -- 
>> ========================
>> Scott Kolbe
>> Postgraduate Student
>> Neuroimaging Group
>> Howard Florey Institute &
>> Centre for Neuroscience
>> University of Melbourne
>> VIC, Australia, 3010.
>>
>> ph:       +61 3 8344 1887
>> email:    [log in to unmask]
>> website:  http://www.neuroimaging.org.au/index.php?id=383
>>


-- 
========================
Scott Kolbe
Postgraduate Student
Neuroimaging Group
Howard Florey Institute &
Centre for Neuroscience
University of Melbourne
VIC, Australia, 3010.

ph:       +61 3 8344 1887