Print

Print


Lythgoe, David  Mark Jenkinson thank you very much.

Shiliang


2008/4/16, Mark Jenkinson <[log in to unmask]>:
>
> Dear Shiliang,
>
> You are right that there are many ways of dealing with signal loss at the
> acquisition level.
> There is no "software" that can "solve" it at the post-processing level
> though.  We have
> some things implemented to try and reduce the effects of signal loss on
> things like
> registration, but they never restore signal.
>
> So what you need to do is explore the various acquisition options for
> reducing signal loss.
> Thinner slices can help a lot, and so can slice-dependent z-shimming, to a
> lesser extent, if
> you can get it implemented.  In order to cope with increasing TR you could
> consider trying
> parallel acquisition acceleration to get your TR down again, although it
> comes at the price of
> some SNR reduction.  Still, it sounds like you have found most of the
> appropriate methods.
>
> I just wanted to say that I completely agree with David's earlier reply,
> and to clarify that this
> is really an acquisition issue and not one that FSL can deal with in
> post-processing.
>
> All the best,
>        Mark
>
>
>
> On 15 Apr 2008, at 14:28, qiang wang wrote:
>
> > Hi Dave
> > Thank you for your reply. But I found some papers on recovering
> > susceptibility-induced signal losses.
> >
> >  some authors have reported methods for recovering
> > susceptibility-induced magnetic field gradient (SFG)-induced signal losses.
> > The simplest method uses thinner slices to reduce the field change across a
> > slice (1). However, this method reduces both the signal-to-noise ratio (SNR)
> > and the spatial coverage per unit of time. Another method (2每4) effectively
> > compensates for field inhomogeneities
> > using multiple refocusing gradients. However, in practice, many
> > repetitions are needed to sum up to a uniform image. This disadvantage
> > limits its value in fMRI experiments. Some methods using high-order field
> > compensation (5,6) and a two-shot technique (7,8) were more efficient but
> > the repetition time (TR) was still doubled.
> >
> >
> > 1. Young IR, Cox IJ, Bryant DJ, Bydder GM. The benefits of increasing
> > spatial resolution as a means of reducing artifacts due to field
> > inhomogeneities. Magn Reson Imaging 1988;6:585每590.
> > 2. Frahm J, Merboldt KD, Hanicke W. Direct FLASH MR imaging of magnetic
> > field inhomogeneities by gradient compensation. Magn Reson Med
> > 1988;6:474每480.
> > 3. Ordidge RJ, Deniau JC, Knight RA, Helpern JA. Assessment of relative
> > brain iron concentrations using T2-weighted and T2*- weighted MRI at 3
> > Tesla. Magn Reson Med 1994;32:335每341.
> > 4. Constable RT. Functional MR imaging using gradient-echo echoplanar
> > imaging in the presence of large static field inhomogeneities. J Magn Reson
> > Imaging 1995;5:746每752.
> > 5. Cho ZM, Ro YM. Reduction of susceptibility artifact in gradientecho
> > imaging. Magn Reson Med 1992;23:193每196.
> > 6. Glover G, Lai S. Reduction of susceptibility effects in BOLD fMRI
> > using tailored RF pulses. In: Proceedings of the 6th Annual Meeting of
> > ISMRM, Sydney, Australia, 1998. p. 298.
> > 7. Chen NK, Wyrwicz AM. Removal of intravoxel dephasing artifact in
> > gradient-echo images using a field-map based RF refocusing technique. Magn
> > Reson Med 1999;42:807每812.
> > 8. Mao J, Song AW. Intravoxel rephasing of spins dephased by
> > susceptibility effect for EPI sequences. In: Proceedings of the 7th Annual
> > Meeting of ISMRM, Philadelphia, PA, USA, 1999. p. 1982.
> >
> > So I think there are many people do this work. Am I right?
> > Shiliang
> >
> >
> > 2008/4/14, Lythgoe, David <[log in to unmask]>: Hi Shiliang,
> >
> > The best software to use is the pulse sequence on your scanner. If you
> > reduce the slice thickness, you'll reduce the through-plane de-phasing that
> > causes most of the signal loss. If you increase the matrix size, you'll also
> > reduce signal losses due to in-plane echo-shifting.
> > If that solution isn't available, sigloss in FSL can provide an estimate
> > of the signal loss, which can then be used to "correct" the signal. You'll
> > also need a field map.  One problem with this is it locally increases the
> > noise. Also, in some regions the signal loss is so severe, any correction
> > would be completely meaningless.  Finally, I haven't used sigloss directly
> > for some time, but I found it underestimates susceptibility-induced signal
> > losses. In fact, I would advise against attempting to correct for lost
> > signal using post-processing. Others may disagree.
> >
> >
> > Dave
> >
> >
> >
> >
> >
> > -----Original Message-----
> > From: FSL - FMRIB's Software Library on behalf of qiang wang
> > Sent: Mon 14/04/2008 15:03
> > To: [log in to unmask]
> > Subject: [FSL] Which software can solve the problem of
> > susceptibility-induced signal losses in EPI images
> >
> > Dear every one
> >
> > I want to solve the susceptibility-induced signal losses in EPI images.
> > Who
> > can tell me which software can solve this problem? Thank you.
> >
> > Best wishes
> >
> > Shiliang
> >
> >