Nasim:
 
The true noise can be calculated by measuring noise in the background, but you have to be careful.  Because MRI intensities are magnitudes, noisy signals are described by Rician statistics, which were described by S. O. Rice in a 1944 technical paper.  In the absence of any true signal, the statistics are also called the Rayleigh distribution.  If SIG is the true SD of the signal in each quadrature channel, the average measured signal intensity of pure noise is SIG*sqrt(pi/2) ~ 1.253*SIG, and the measured SD of the signal intensity of pure noise is SIG*sqrt[2-(pi/2)] ~ 0.655*SIG.  Therefore, you have to do one of the following calculations:
 
1) Measure the average value in a ROI where there is no image artifact (ghosting, etc.), and divide by sqrt(pi/2) ~ 1.253.
2) Measure the standard deviation (SD) in the ROI, and divide by sqrt[2-(pi/2)] ~ 0.655.
 
For example, in a DTI paper that should appear soon in Magn Reson Med, we found a mean value of 24.1 and a SD of 17.2, yielding estimated noise levels of 24.1/1.253 = 19.2 and 17.2/0.655 = 26.3.  The two numbers will rarely agree exactly.  Ideally, you would report both noise calculations.  If you want to report a single number, I suggest taking the average of these two calculations, unless you have reason to believe that one is more accurate than the other.
 
If you draw a ROI in tissue rather than noise, you add tissue heterogeneity to the true noise.  If your tissue noise is less than or equal to your background noise, it is safe to use it.  If the tissue noise is greater than the background noise, I suggest using the background noise.
 
The following references provide more information.
 

Edelstein WA, Bottomley PA, Pfeifer LM. A signal-to-noise calibration procedure for NMR imaging systems.  Med Phys 1984;11:180-185.

Henkelman RM. Measurement of signal intensities in the presence of noise in MR images.  Med Phys 1985;12:232-233.  [Published correction: Med Phys 1986;13:544]

Bernstein MA, Thomasson DM, Perman WH. Improved detectability in low signal-to-noise ratio magnetic resonance images by means of a phase-corrected real reconstruction.  Med Phys 1989;16:813-817.

McGibney G, Smith MR. An unbiased signal-to-noise ratio measure for magnetic resonance images.  Med Phys 1993;20:1077-1078.

Gudbjartsson H, Patz S. The Rician distribution of noisy MRI data.  Magn Reson Med 1995;34:910-914. [Letter, related letter, and response: Magn Reson Med 1996;36:331-333]

Dietrich O, Heiland S, Sartor K.  Noise correction for the exact determination of apparent diffusion coefficients at low SNR.  Magn Reson Med 2001;45:448-453.

 

--
Peter B. Kingsley, MRI Physicist        [log in to unmask]
Department of Radiology / MRI
North Shore University Hospital
300 Community Drive, Manhasset, NY 11030, USA
Tel (516) 562-2842       Fax (516) 562-3561

-----Original Message-----
From: SPM (Statistical Parametric Mapping) [mailto:[log in to unmask]]On Behalf Of Nasim
Sent: Thursday, March 24, 2005 10:25 AM
To: [log in to unmask]
Subject: [SPM] Signal to Noise Ratio

Hi all,

 

I would like to know is there any consensus on how to define noise when you are doing Region of Interest (ROI) based analysis and you want to determine noise so that you could define SNR in a ROI?

 

In some papers I see they define a ROI in the background and measure noise as the mean of that ROI and consider it as noise value every where in the image. On the other hand in some other papers they use the standard deviation of the values on the desired ROI as noise.

 

Thanks

Nasim


Do you Yahoo!?
Yahoo! Small Business - Try our new resources site!

_____________________________________________________________________


The information contained in this electronic e-mail transmission

and any attachments are intended only for the use of the individual

or entity to whom or to which it is addressed, and may contain

information that is privileged, confidential and exempt from

disclosure under applicable law. If the reader of this communication

is not the intended recipient, or the employee or agent responsible

for delivering this communication to the intended recipient, you

are hereby notified that any dissemination, distribution, copying

or disclosure of this communication and any attachment is strictly

prohibited. If you have received this transmission in error, please

notify the sender immediately by telephone and electronic mail,

and delete the original communication and any attachment from any

computer, server or other electronic recording or storage device

or medium. Receipt by anyone other than the intended recipient is

not a waiver of any attorney-client, physician-patient or other

privilege. Thank you.