Print

Print


From your description it sounds as if your spirometers may use a  range of
different measurement techniques.

Some, such as the "Gold standard" traditional Vitalograph, measure volume
directly, while others use pneumotachographs (flowheads), rotating devices or
orifice plate systems to measure flow rate and then integrate that to get
volume.

Pneumotachographs generally have a linear flow/pressure  response but have a
finite resistance to airflow, which should  be below a certain value to be
acceptable in this application, and are  sensitive to contamination. They tend to
follow the flow signal quite  well.

Rotary devices also measure air flow but can suffer from overshoot when the
flow rate changes quickly or inertia when the rate of change is slow. They
tend  to have lower resistance to airflow than pneumotachographs.

Orifice plate devices have very little resistance to airflow and will not
suffer much from contamination but need electronic correction of the
flow/pressure response and will not be very stable at low flow rates.

Each technique offers a different resistance to the person using the
equipment, which might affect your readings, and offers a slightly  different
response depending on the mix of rapidly or slowly changing flow  rates.

My guess would be that when you use your syringe to check calibration, you
will apply it to each device in a way that will produce a sudden increase in
flow which will plateau and continue for a couple of seconds and which will
terminate suddenly when you hit the end stop. Your devices are calibrated to
this step type of flow signal.

This is not however what a person does when using the equipment. There  is an
explosive initial burst which immediately tails off and might continue for  a
few seconds at a very small flow level before terminating.

You could try to simulate that with a syringe if you want to assess which,
in the real world, gives consistently best results.

There are other factors which could be at play of course but these are some
of the issues which manufacturers have to try and deal with when designing
equipment of this type.

You could try P&A   tel 01204 676180   as an  independent testing and
calibration company.


From:-
G M Instruments Ltd
Unit 6, Ashgrove
Ashgrove  Road
Kilwinning
KA13 6PU
Scotland
UK

Tel + 44 (0)1294  554664
Fax + 44 (0)1294  551154
WWW.GM-INSTRUMENTS.COM

~~~~~~~~~~~~~~~
Please remove this footer before replying.