I recently collaborated with some neonatologist
colleagues by comparing neonatal blood glucose measurements using the Bayer
Rapidlab 860 blood gas analyser glucose electrode with those obtained using the
YSI 2300 Stat and the Vitros 750 XRC.
As part of this study, we considered the technology
of the glucose electrode (with assistance from Bayer).
Although capillary whole blood samples were
presented to the analyser, the electrode membrane is impermeable to protein (and
red cells) so only the plasma water is in contact with the electrode.
As the glucose is measured in a fixed volume of solvent (temperature
is constant) it is close to being a plasma glucose molality
measurement.
The calibration is also crucial, the Bayer method
is calibrated against the hexokinase/glucose-6-phosphate
dehydrogenase plasma glucose reference method.
The end result is that there was extremely
close agreement between the "whole blood" glucose electrode measurements and the
plasma glucose method (Vitros 750). However there was a small
discrepancy between the Bayer results and those for the YSI using whole blood,
which was consistent with the difference in water content between plasma and
whole blood.
We also looked at the effect of haematocrit and
could find no effect at all between 20 and 80%, which is presumably a
consequence of the membrane retarding
the haemoglobin.
In short, certainly for this electrode, the results
using whole blood were extremely close to those obtained using the
laboratory plasma glucose method. This work has been accepted by the
Annals, although we only looked at samples from neonates with low glucose
concentrations.