Hi Matt,
Where are you measuring your baselines?
Paul
>Hi All,
>
>We have recently changed our multiplier from a Johnston to a Blazers
>and due to this have higher than normal backgrounds as the machine
>recovers from being vented. Despite fairly high backgrounds, we
>proceeded to run some Fish Canyon sanidine to see where we were at
>with respect to reproducibility. We consistently obtained slightly
>less radiogenic yields, observed a slight age variation that was
>correlated to signal size and MSWD values for 6 analyses of about 2
>to 3.
>
>We determined that we could subtract an additional absolute value of
>about 4e-18 moles from mass 36 from our measured value and thus
>increase our radiogenic yield values and obtain populations with
>MSWD's of 1 or less. We refer to this as a phantom 36 correction.
>
>In an attempt to understand why our apparent 36 background was less
>during a blank than during an analysis, be began looking at mass 37
>and mass 35 with a little more rigor. We found that for an air
>analysis or for an unirradiated sanidine we consistently measured
>negative 37 and 35 following blank correction. That is, we had less
>35 and 37 during analysis than during a blank run. In fact, running
>any sample of sanidine, air, biotite or amphibole always returns at
>negative 35 signal of about 4e-18 moles.
>
>We believe that it is no coincidence that the negative 35 signal
>matches in magnitude the correction that we need to make to mass 36
>in order to get quality Fish Canyon sanidine populations. We
>hypothesize that introduction of hydrogen from a sample or air split
>is reacting with our chlorine backgrounds in the mass spec such that
>35 and 37 are reduced during analysis (compared to a blank) and in
>turn we create HCl (thus the mass 36 interference) as well as minor
>mass 38 (37Cl+H).
>
>Monitoring mass 2 during gas introduction shows that our getters
>come to a consistent steady-state value that is non zero (3e-15
>moles) no matter what we put in the instrument. Much of the H is
>pumped out by the ion pump after analysis. Also, closing the mass
>spec getter with the machine static shows a pretty dramatic hydrogen
>degassing of the mass spec that is readily pumped upon opening the
>getter. Thus, our getters seem to be working normally.
>
>Our present Mass spec backgrounds are (moles):
>
>40 - 2e-18
>39 - 5e-19
>38 - 5e-19
>37 - 8e-18
>36 - 2.5e-18
>35 - 2.5e-17
>
>As you can see, 35, 36, and 37 are higher than we'd like, but not
>outrageously bad. We believe we have not observed this prior to
>venting the machine because our backgrounds were in the E-19 range
>and thus any introduction of hydrogen had nothing to react with to
>cause our HCl interference.
>
>We would be curious to know if others have seen any similar behavior
>to what is described above. If any one has at present the
>unfortunate circumstance of relatively high backgrounds, we would
>appreciate information on analysis of unirradiated samples (i.e.
>negative 37) or any other runs to test for negative 35. Also, anyone
>willing to take a look at hydrogen could help us evaluate the
>quality of our getters. We think we have exhausted any electronic
>issue that could cause this apparent behavior (i.e. decay of signal
>following measurement of a large beam, signal non-linearity, etc.).
>The chemistry of our problem seems to make sense. Any feedback would
>be welcomed.
>
>Thanks,
>
>Matt
>
>
>
>
>--
>Matt Heizler
>NM Bureau of Geology and Mineral Resources
>NM Tech
>801 Leroy Place
>Socorro, NM 87801
>
>Office 505-835-5244
>Argon Lab 505-835-5271
>Main Office 505-835-5420
>FAX 505-835-6333
>http://geoinfo.nmt.edu/staff/mheizler/home.html
|